I'm hoping theres someone here that can give me some advice. I have a seaward combi compact tester which i put in for calibration. I have been told there is a fault on the tester as when it runs an earth fault loop impedance test it only ever records 0.67 ohms. I found this very strange as i have been using the tester regularly and have numerous installation certificates on which none of the recorded values are 0.67ohms. They actually vary from 0.15 to 1.5ohms. I also know they are approximately correct as i also use the calculation method for Zs.
I was rather concerened when the man testing my machine asked if i could take the manual with me as he had no documentation on it. Yet they are supposed to be seaward approved. He didn't even know how to set the machine to carry out the test.
I have just picked the tester up and refused to pay the bill (£75) as in my opinion they are not competant to have been calibrating the machine.
I have the calibration certificate which states on 7 different applied test values for L-PE (0.57-100ohms) the result was 0.67ohms on all 6 then 1ohm result for 100ohm applied.
The strange thing is i have just run the loop test across my PME supply at home (straight across the tails) to find Ze and got a value of 0.18ohms which i would think is about right. I then borrowed a friends robin loop tester and that recorded 0.19ohms.
Is it possible they wern't testing my machine properly? Can anyone suggest a way in which i can check the EFLI readings myself.
I already check the continuity meter and ins resistance with resistors. But cannot check the EFLI or RCD trip times myself.
I was rather concerened when the man testing my machine asked if i could take the manual with me as he had no documentation on it. Yet they are supposed to be seaward approved. He didn't even know how to set the machine to carry out the test.
I have just picked the tester up and refused to pay the bill (£75) as in my opinion they are not competant to have been calibrating the machine.
I have the calibration certificate which states on 7 different applied test values for L-PE (0.57-100ohms) the result was 0.67ohms on all 6 then 1ohm result for 100ohm applied.
The strange thing is i have just run the loop test across my PME supply at home (straight across the tails) to find Ze and got a value of 0.18ohms which i would think is about right. I then borrowed a friends robin loop tester and that recorded 0.19ohms.
Is it possible they wern't testing my machine properly? Can anyone suggest a way in which i can check the EFLI readings myself.
I already check the continuity meter and ins resistance with resistors. But cannot check the EFLI or RCD trip times myself.